My opinion is that you are seeing nothing but nickel. You apparently processed without inquartation. Do follow the wise advice from butcher and test your stannous chloride against a standard solution to insure that you truly have eliminated all values.
An interesting testing procedure for you, to expand your testing skills; a test with DMG will be conclusive about an absence of palladium, which I do not suspect. It would have turned your solution quite dark brown, but on the outside chance you have traces, DMG will yield a yellow precipitate, and it's very sensitive, displaying the yellow reaction when very little is present.
After testing for palladium with DMG, if you'd like to see a reaction for nickel, place a drop of your solution in a spot plate cavity, then add a drop of ammonium hydroxide. A drop of DMG will then yield a beautiful pink reaction. If your solution contained copper instead of nickel, you would not get a pink reaction with DMG, and the addition of ammonium hydroxide would have yielded a brilliant blue reaction.
If you are clueless about DMG, Hoke is your friend.
Harold